Ludlum model 3 issues
Sarmad suseyn
Real name: Sarmad
Post by Sarmad suseyn »
Hi
I just bought ludlum model 3 counter. It counts good with the
Probe 44.3
The issue is that the high voltage screw is not adjusting the voltage. The only voltage that i can get is 750 v.
When i opened the box i noticed : see pictures
1-There is two wires were cut(brown and red)
2- the high voltage points gives 750,900,1600,2000 volts
3- a white wire is should it be there?
Please help .
If any one have the same model can you send pictures of his own
Many thanks
Re: Ludlum model 3 issues
See last photo..Easy to answer. Inline plug seen on bottom of board has 8 points. count them. cable is 10 conductor. Two wires are superfluous and just cut off. No problem.
Post by Rex Allers »
Richard, not sure your easy answer is correct.
I've never had a Ludlum Model 3 so have zero direct knowledge but here's a version of that pic that I modified.
So I added two outlines of possible connections on the board, one is orange, one is green. You propose an inline set of connection nodes for the cable. In my picture that would be the section outlined in orange. Yep, 8 pins, hence 8 wires.
Now look at the section I outlined in green. To me it seems to be more aligned with the cable and has two rows of 5 nodes. 10-pins for the 10 wires in the flat cable. Flat cables often terminate in two-row connectors. So I think the two cut wires do have pins into the board, but are they used? Dunno.
I haven't taken any more time than looking at this pic, but Ludlum does share a lot of info. The manuals usually have schematics. I checked and there is a model 3 manual with schematics here...
https://ludlums.com/documents/product-manuals
There could be version issues, but with some studying a motivated student might find if those two wires should go somewhere.
One other observation. The soldering job for the blue wire at the bottom of the pic between two front panel terminals looks pretty poor. I would take this as a hint that someone has hacked at this device for some reason.
Sarmad suseyn
Real name: Sarmad
Thanks for the help guys
In defense of my original point.
Sarmad suseyn
Real name: Sarmad
Hi Richard
Thanks for the picture. My only concern is that i can't regulate
The high voltage screw. Its stuck on 750v . Mean while the high voltage ports are giving 900 ,1500, 2000 v. Tried to change the variable resistor without luck. Any wonderful ideas?
Thanks for your time
Without seeing the thing or having a schematic, I have no clue.
Sarmad,
I spent a bit of time today trying to understand the Model 3 circuits. It seems that the Mod 3 manual on the Ludlum page I linked earlier has schematics and board layouts that don't match yours. I assume the doc is for a newer version.
One significant difference is the shape of the board that has the row of adjustment pots on it. Looking at other manuals I found boards with the same basic shape in manuals for other '3' models (3A, 3-97, 3-98). These shouldn't be the same, though, because they have more than 5 adjustment pots. So similar but not the same.
In these other manuals, the main board has a similar shape but not the same component layout. The main board is the one where you showed the voltage steps.
To really understand what is up with your unit, you need to get the proper Model 3 manual with schematic section that matches yours. If you contact Ludlum with your model and serial number you may be able to get the proper manual. My attemps at web searches today didn't find any documentation that seemed to match your boards.
I do think that your suspicion about the white wire may be valid, as it may be a mod that was added for some reason. It does seem to me that if the HV adj signal (from the HV pot to the main board) is unmodified then the HV should adjust through some range even if the white jumper has been added.
If you can find the proper documentation for your model, I am willing to try to help you figure out the details about what could be going on.
Oh, and I looked at the documentation for the 3A that has a board shaped like the one in your pics (with the adjustment pots). Not the same circuit but similar, so the connections in the area that I outlined in green in an earlier post may actually be two connectors, one 5-pin and one 4-pin.
A guess from me. Assuming the top adjustment pot on the front panel is HV, then I would guess the signal at the purple dot in the following pic would be a low dc voltage (maybe 0 to 1.2V) that changes as the HV pot is turned. That signal should connect directly to some input pin on the main board. The HV out should change proportionally with that voltage. If you have a DVM maybe you can check out voltage and connection to the main board.
Good luck.
Sarmad suseyn
Real name: Sarmad
Dear Rex Allers
Thanks a lot. Now i can adjust it.
Did you find something that made it start working? Does it go beyond 750 V?
I'm with Rex......."Thanks it works now", is not a fix! What fixed it!!!! We've gone through this posting for a bit now.... What fixed it??? We were left hanging by your response or lack thereof.
Sarmad suseyn
Real name: Sarmad
Hi Richard
I think my problem was the variable resistor
It is 64 watt resistor. And when i checked the ohms . It gives zero
Value
I think this was the problem. Check it with a 100ohm resister
It burned my multimeter 1000v .
Cant change it right now because of curfew . Every electronics shop is closed. But im sure this is the problem
Thanks Richard
Sarmad,
None of that explanation makes any sense to me. There's no 64W pot in a Ludlum 3. The batteries probably couldn't supply 64W if they were shorted. The little pot for HV adjust setting is probably 100K ohm. In the circuit it should draw microamps and micro watts.
Not sure how you were measuring resistance but 0 ohms (a short) seems very unlikely. I can't think of how a single 100 ohm resistor could be used in your diagnosis. I also can't imagine what in a Ludlum 3 could burn your meter.
Of the schematics I have found, the Model 3A looks probably closest to being similar to the pictures you have provided. Not the same but I would guess the circuits are very similar. I edited those schematics to make one page with the HV supply circuit and the HV adj pot as it would connect. I also added a purple dot in the same place where I put one in an earlier version of your picture that I posted. If you can't understand that circuit, you are probably not in a good position to diagnose your device.
