First Neutrons in Fusor II
- Real name: Jack Rosky
Written Overview:
Neutron Detection- Our neutron detection consisted of a bubble dosimeter rated at 29 bubbles/mrem. Our He-3 counter will be running shortly after we get it calibrated against a proper source.
Vacuum system- Our vacuum system consists of the same 1 CFM mechanical pump (Pfiefer-Balzers). Despite being small, it easily pumped our entire system down to 6.5 mTorr before turning on the diffusion pump. Our diffusion pump is a 3” Varian Air-Cooled pump with a water cooled baffle (separate from the pump) in order to prevent the oil from backstreaming. After our diffusion pump is a CF 2.75” bellows valve in order to throttle down the system. Our main vacuum chamber is a 9” inner diameter sphere.
Gas System- Our gas system consists of a 10 liter lecture bottle of 99.8% deuterium, a ball valve, and a 0-10 sccm Mass Flow Controller. Our gas system provides very precise control over gas flow into the chamber and allows us to maintain a constant pressure. Because of this, we were able to run the fusor, hands-free, for a total of 45 minutes before we had to stop due to the HVPS arcing somewhere in our HV line.
High Voltage- Our high voltage system consists of -40kv 7.5mA Glassman power supply. This supply required us to build our own controller to supply a 0-10 reference voltage to the supply. The panels on our controller are not fully functional, so instead we use a fluke high voltage probe that records on a 2000: 1 ratio. This is why when read at -10 volts the supply is actually giving out -20kv.
Data collected from run:
Date and Time of run: August 10, 2016 at 1:15 PM
Operators: Scott Moroch and Jack Rosky
Voltage: 22kv
Current: Currently unknown. Assumed to be about 4mA
Pressure:17mTorr
Grid Material: Stainless Steel
Grid type: 3- ring. Rosenstiel method
Background count on detector: 0 bubbles for 1 hour background check
Bubbles per exposure time: 12 bubbles/ 20 minutes
Distance from center of grid: 6”
Total length of run: 55 minutes
Total time run without operating controls: 45 minutes
Brief overview of fusor subsystems during first fusion run:
https://www.youtube.com/watch?v=IKUCQfc9zpY
If I have seen further than others, it is by standing upon the shoulders of giants.
-Issac Newton
-Issac Newton
- Steven Sesselmann
Re: First Neutrons in Fusor II
Jack,
Congratulations on completing a really nice clean fusor project, and achieving good neutron numbers at only 22 kV.
I plugged your data into the fusion calculator and it all looks like a good result.
Thumbs up from me..
http://www.gammaspectacular.com/fusion_calculator.html
Thumbs up from me..
Good Job, Scott and Jack! Your names are added to the neutron club listing.
Great work and good reporting! Your efforts demonstrate good experimental procedures. Congratulations.
Hello All!
First I'd like to say thank you for all the nice words and congratulations. I would just like to point out that this is not Scott and I's first fusion. It is simply our first fusion in our new system. Our first fusion was done around this time last year. Here is the neutron club claim from last year: viewtopic.php?f=18&t=10277
Thank you Richard for admitting us to the neutron club, but we happen to already be there. Sorry for the confusion from the title having "first neutrons" in it.
We hope to bring up the voltage soon as we are fixing a place where it arcs past 25kv. Stay tuned for increased fusion numbers and for our experiment regarding different metal plates within the fusor.
Thanks again,
Jack
Thanks again,
Jack
-Issac Newton
-Issac Newton
- Scott Moroch
- Real name: Scott Moroch
Today Jack and I saw an increased amount of fusion with our best run being 1.19e5 n/s isotropic (10 bubbles in 164 seconds on a 29bubble/mrem BTI at a distance of 5 inches). We also managed to capture some nicer star mode pictures which are attached before. We are hoping to continue running our system until we hit 1e6 n/s in order to conduct activation experiments.
Scott Moroch
Scott Moroch
-Albert Einstein
-Albert Einstein
Fabulous pictures! That is the look of fusion! Back in the days when we made full geodesics, the rays were fabulous. I got as many as 14 star rays in one of my early images.
Sorry about the double posting. I irradicated the latest admission statement. Thanks for the heads up.
- Scott Moroch
Our latest fusion run hit 180,000 n/s at 38kV, 3mA and 11mTorr. This run was conducted after taking the previous grid out in order to change the grid design slightly. Pictures of star mode are below. You will now notice a short length of stainless steel tubing emerging from the top of the ceramic before the grid itself.
Scott Moroch
Scott Moroch
-Albert Einstein
-Albert Einstein
Really nice photo's of your star mode.
It is so visibly obvious and so believable your numbers from carefully executed fusor work. I wish all such claims of fusion were so obvious, visually and experimentally reported. Keep going and you will hit 500K soon, easy.
You might be able, right now, to do silver activation and check it with a simple GM counter using good technique. (background count of 5 minutes - no fusor. Silver count within 10 seconds of fusor shutdown of one minute - remove silver and do a final background of 5 minutes.) Average the CPM of both background runs and compare with the 1 minute silver measurement.