Hydrogen leak detection

Post by Ross Moffett »

I wonder if hydrogen might be useful for leak detection as helium is?

Hydrogen detector with 100 ppm+ sensitivity at $8 is hard to compete with, looking at pricing for used leak detectors.
User avatar
Carl Willis
Posts: 2841
Joined: Thu Jul 26, 2001 7:33 pm
Real name: Carl Willis
Location: Albuquerque, New Mexico, USA
Contact:

Re: Hydrogen leak detection

Post by Carl Willis »

Hydrogen is a major component of the residual gases in technical vacuum work, so wouldn't be high on the list as a leak detection tracer for vacuum systems. At the gross leak level, however, almost any common gas can be used as a leak tracer with an appropriate detector.

-Carl
Carl Willis
http://carlwillis.wordpress.com/
TEL: +1-505-412-3277
User avatar
Monroe Lee King Jr
Posts: 105
Joined: Thu Aug 22, 2013 11:57 pm
Real name: Monroe Lee King Jr
Location: Lampasas, Texas
Contact:

Re: Hydrogen leak detection

Post by Monroe Lee King Jr »

I would think a Pirani gauge and Helium would be better. The gauge would change a detectable amount for a small leak. It would take a good deal of time to do it properly I would imagine but if you have ever searched for a tiny vacuum leak before you already know that.

Monroe
Today's the Day! We go into Space!
Ross Moffett
Posts: 89
Joined: Mon Aug 26, 2013 10:24 pm
Real name: Ross Moffett

Re: Hydrogen leak detection

Post by Ross Moffett »

My objective is to have some way to evaluate my vacuum chamber welds, so per-contamination shouldn't be an issue. I was under the impression that helium would be used due to its better tendency to escape through small gaps (a "spongy" weld, for example). Would a refrigerant leak detector, or other leak detector, really work just as well for evaluating high-vacuum welds?

Maybe someone knows, I'm also concerned that the pre-existing weld on my stainless welded "garden sphere" may be leaky. Is there any experience with this, maybe the reason most people pursue hemispheres welded to conflat?
User avatar
Monroe Lee King Jr
Posts: 105
Joined: Thu Aug 22, 2013 11:57 pm
Real name: Monroe Lee King Jr
Location: Lampasas, Texas
Contact:

Re: Hydrogen leak detection

Post by Monroe Lee King Jr »

You might be able to Zyglo test for cracks it's not that expensive. You can make your own test with talcum powder and penetrating oil if you powder one side and spray penetrating oil on the other cracks show up. If you've made an X-Ray device this is a good place to use it. Brazing might be better than welding the thin stuff. Unless of course you are an experienced TIG welder. Magnaflux works on some stainless steels. Pressurizing a garden bowl with iffy welds could be dangerous indeed. You can use a Pirani gauge and test under vacuum. You just have to control your helium use, use a discarded ink pen attached to some surgical tube and dose a small area. Then wait for that to dissipate and give time for the Pirani gauge to register. Move on to the next little area pinch off the helium supply as fast as you can every time to keep the atmosphere as free of helium as you can. If you go releasing a large amount of helium you'll get a false indication from some other area that picks it up. Once you find a spot test it several times to make sure you found the spot. Freon would work with that gauge too, but don't inhale and gas from the chamber because freon makes phosgene gas if exposed to a hot filament (or cigarette) ect... a deadly gas.

Monroe
Today's the Day! We go into Space!
User avatar
Carl Willis
Posts: 2841
Joined: Thu Jul 26, 2001 7:33 pm
Real name: Carl Willis
Location: Albuquerque, New Mexico, USA
Contact:

Re: Hydrogen leak detection

Post by Carl Willis »

Hi Ross,

Helium leak detection is the standard, and most sensitive, method used with vacuum systems. One of the advantages of buying commercial vacuum components is that all welds have been helium leak-checked. Any machine shop that specializes in vacuum technology will have a helium leak detector and trained operators. University and industry research labs typically have a helium leak detector. In home laboratories, they are uncommon due to their cost. You'll find no reference to any fusioneers owning these machines, although a few people do have RGAs that can be operated as helium (or other gas) leak detectors.

A frequently-documented method of hunting leaks in fusors is to start a discharge in the fusor and then spray or wipe suspect areas with alcohol or some other hydrocarbon. Leaks cause the discharge to change color from the pink hues typical of air or the brick red of deuterium, to a characteristic dark blue.

Vacuum gauges respond somewhat unpredictably to the ingress of tracers into a leak. A volatile liquid such as water or alcohol may initially cause a pressure surge when it enters a leak channel, but frequently these liquids actually plug the leak through surface tension and / or freezing.

-Carl
Carl Willis
http://carlwillis.wordpress.com/
TEL: +1-505-412-3277
Ross Moffett
Posts: 89
Joined: Mon Aug 26, 2013 10:24 pm
Real name: Ross Moffett

Re: Hydrogen leak detection

Post by Ross Moffett »

Any comments about what I might expect to find with the pre-welded sphere?

I'm wondering if it will be a good idea to cut out the weld with a lathe and re-do it to CF flanges.. but it's not really necessary so far as fusing and vapor deposition (the other thing I will be doing in my chamber) are concerned, since I could insert the samples/fixtures through smaller flanges. I'd like to not cut and weld to large diameter conflat if previous fusioneers have been down this path and found the factory welds satisfactory. Mine is machine welded, there is no "stack of dimes" effect, and the weld appears to be very uniform and good.
User avatar
Richard Hull
Moderator
Posts: 14992
Joined: Fri Jun 15, 2001 9:44 am
Real name: Richard Hull

Re: Hydrogen leak detection

Post by Richard Hull »

Nothing whatsoever will be known about the vacuum readiness of any weldment until a quality, calibrated vacuum gauge capable of reading to 10e-6 torr is in hand and the weldment is tested under a vacuum of at least 10e-5 torr.

Richard Hull
Progress may have been a good thing once, but it just went on too long. - Yogi Berra
Fusion is the energy of the future....and it always will be
The more complex the idea put forward by the poor amateur, the more likely it will never see embodiment
Ross Moffett
Posts: 89
Joined: Mon Aug 26, 2013 10:24 pm
Real name: Ross Moffett

Re: Hydrogen leak detection

Post by Ross Moffett »

Then that is what I shall do. I'm going to weld a 2.75" Conflat to it and try both methods.. I figure I can stick this hydrogen detector in the exhaust of my roughing pump while I blow hydrogen over the welds. At $8 I can breadboard something to see if it works, will update when (if? :) ) such time comes.
John Futter
Posts: 1848
Joined: Wed Apr 21, 2004 10:29 pm
Real name: John Futter
Contact:

Re: Hydrogen leak detection

Post by John Futter »

Ross for your detector to work you would have to have a gross leak
A leak that could be found by pressurising the sphere and looking for bubbles with soapy water.
Helium leak detectors are 10,000 to 100,000 times more sensitive than that detector
Ross Moffett
Posts: 89
Joined: Mon Aug 26, 2013 10:24 pm
Real name: Ross Moffett

Re: Hydrogen leak detection

Post by Ross Moffett »

I looked up the high end sensitivity of a helium leak detector (1 cm^3 per century rate!) and you're right.. Do you happen to know what these things use as the detection method? Based on a patent search, a few patents may have those really expensive thermal tubes used in industrial gas analyzers. It seems like the portable ones are more likely to be using spectrometry. Helium has some emission peaks at 667.8nm and ~710nm. Hydrogen has a peak at 656.2nm, which should also be easy/cheap to catch with a spectral analyzer. One of the patents in particular just looked for a rapid change, since it wasn't really trying to nail down the exact amount of leakage but just the fact that there was one.

I bet this can be done really cheaply with an LED in the just-right spectrum, a photodetector, a filter and a basic op-amp circuit. Scientific grade filters can run ~$100, so maybe a DIY spectrometer would be better. It could double as a He detector for fusion.

These people lay out how the typical DIY spectrometer works, though there are several ways to skin that cat.
http://publiclab.org/wiki/spectrometer
John Futter
Posts: 1848
Joined: Wed Apr 21, 2004 10:29 pm
Real name: John Futter
Contact:

Re: Hydrogen leak detection

Post by John Futter »

Ross
every Helium leak detector I have seen that is any good is an ion mass spectrometer tuned to the helium mass unit.
some of the more expensive models can tune to other gases such as argon hydrogen etc.
it is because of this that they can detect down to a few helium atoms ie multiples of 1.6 x 10^-19 amps.

for bigger leaks both the penning and pirani gauges react to hydrocarbons (alcohol acetone etc) and give a kick to the indicated pressure --this is useful when the leak is too large for the helium leak detector to be useful.

by putting an RGA on a system it can be used as a helium leak detector and most moern RGA's offer this in the supplied software
User avatar
Richard Hull
Moderator
Posts: 14992
Joined: Fri Jun 15, 2001 9:44 am
Real name: Richard Hull

Re: Hydrogen leak detection

Post by Richard Hull »

Helium is used as it is monatomic and one of the tiniest of all atoms. It can creep into crevices much smaller than the diatomic hydrogen molecule with its giant molecular bond. John is right....Almost all helium leaks are simple mass specs tuned for one mass (helium). The amateur fusioneer has virtually zero use for one of these puppies as fusion is easily achieved in arather leaky system that would fail a helium leak test.

Richard Hull
Progress may have been a good thing once, but it just went on too long. - Yogi Berra
Fusion is the energy of the future....and it always will be
The more complex the idea put forward by the poor amateur, the more likely it will never see embodiment
User avatar
Dennis P Brown
Posts: 3159
Joined: Sun May 20, 2012 10:46 am
Real name: Dennis Brown

Re: Hydrogen leak detection

Post by Dennis P Brown »

No one in a lab would use hydrogen for leak detection since the gas is so damn flammable (and the flame is invisible!) and even explosive if there is a build up in the room. Helium can be released in large volumes, safely - hydrogen, well, that would be problematic ... . Releasing hydrogen into a room is just a bad idea in general, I would think ... .
ARDYN VAN FOSAAEN
Posts: 19
Joined: Sat Jul 06, 2013 10:27 am
Real name: ARDYN VAN FOSAAEN

Re: Hydrogen leak detection

Post by ARDYN VAN FOSAAEN »

One way would be to pressurize your vac system with H2 then use the sensor probe along all joints, valves and welds. Vac system 1st to remove all O2 then pressurize sys to a little over atmospheric pressure then test for leaks. When done testing, flush system with nitrogen to remove H2 prior to repair or use.