Spellman noise

This forum is for specialized infomation important to the construction and safe operation of the high voltage electrical supplies and related circuitry needed for fusor operation.
Post Reply
Nicker
Posts: 29
Joined: Tue Jul 21, 2009 12:14 pm
Real name: Nick DePasquale
Contact:

Spellman noise

Post by Nicker »

Testing the power supply from 0--62kv it makes some odd noise from 45-54 -kv. Like a lose serpentine belt. What could this mean?

http://www.youtube.com/watch?v=wSwtp0qHkjg
Dustin
Posts: 105
Joined: Mon Jul 26, 2010 8:40 pm
Real name:

Re: Spellman noise

Post by Dustin »

Its probably just acoustics from the windings vibrating in the fields in a specific load range.
Its fairly common for switchmodes to do this. (annoyingly)
David Geer
Posts: 136
Joined: Wed Nov 24, 2010 7:51 am
Real name:
Location: Colorado Springs, CO

Re: Spellman noise

Post by David Geer »

I agree with Steven. I've experienced this in the past with other pieces of equipment and it's just harmonics in the vibrations AND yes, it's very annoying if you work in that range continuously but it's harmless.
- David Geer
Nicker
Posts: 29
Joined: Tue Jul 21, 2009 12:14 pm
Real name: Nick DePasquale
Contact:

Re: Spellman noise

Post by Nicker »

Thanks for the peace of mind.
Cliff S
Posts: 48
Joined: Fri Nov 21, 2008 1:25 pm
Real name:

Re: Spellman noise

Post by Cliff S »

Nick,
Sorry to spoil your peace of mind...but this is not normal. Sounds like an instability in the voltage regulation loop. Unit may have been damaged at some point in its life.
Often the point(s) of instability will change as a function of line and/or load. Under some small load current, it may disappear. Or, it may get worse. You would probably notice this instability if you were monitoring the HV output on a scope, via a divider or blocking capacitor. (Don't try this unless you are experienced in HV measurement techniques!!!). You can also try to look at the kV test point on the rear panel. It may have adequate BW to see the instability. (Pin 6 on the terminal block or connector). Check our website out for the operation manual of the SL300W series.

http://www.spellmanhv.com/en/Technical- ... SLMAN.ashx

Best Regards,
Cliff Scapellati
Executive Vice President of Engineering
Spellman High Voltage Electronics
631-630-3110