Great American chamber bake-out
Great American chamber bake-out
I think I already know the answer to this but:
Is it worth while to bake out a vacuum chamber with open 6 inch flange (open to atmosphere). The 6 inch flange is intended to be attached to a turbo pump, which I don't want to be exposed to high temperatures. I am not sure If I will be able to attach the pump while the chamber is very hot without burning myself. I also have a vacuum gauge with a 1.33CF flange that has plastic for the electrical connector. That will have to also be connected after the bakeout.
I could close the 6 inch port during the bake out with a flange blank, and perhaps pump down using a roughing pump and back fill with dry argon or nitrogen, and then attach the turbo. Ideally I would use a 6 inch gate valve, but they are extremely pricely and rare on ebay.
Second, can I rough pump down the chamber through the turbo's exhaust port, or do I need to use a port on the vacuum chamber?
Any advice is welcome.
Is it worth while to bake out a vacuum chamber with open 6 inch flange (open to atmosphere). The 6 inch flange is intended to be attached to a turbo pump, which I don't want to be exposed to high temperatures. I am not sure If I will be able to attach the pump while the chamber is very hot without burning myself. I also have a vacuum gauge with a 1.33CF flange that has plastic for the electrical connector. That will have to also be connected after the bakeout.
I could close the 6 inch port during the bake out with a flange blank, and perhaps pump down using a roughing pump and back fill with dry argon or nitrogen, and then attach the turbo. Ideally I would use a 6 inch gate valve, but they are extremely pricely and rare on ebay.
Second, can I rough pump down the chamber through the turbo's exhaust port, or do I need to use a port on the vacuum chamber?
Any advice is welcome.
-
- Site Admin
- Posts: 2124
- Joined: Fri Jun 14, 2002 2:26 pm
- Real name: Frank Sanns
Re: Great American chamber bake-out
Heating a vacuum chamber open to the air will not do a thing for you. It would be a waste of time. The time would be better spent being sure everything on the inside of the chamber was clean and degreased. Connect everything then pump down. You can draw slowly out of the turbo pump if you need to but I don't think it is the wisest because any debris that is left over in the chamber can be sucked into your turbo pump and this is not a good place for it. Once the chamber is pumped down, huge air currents will not sweep large debris into the pump so at high vacuums drawing from the bottom of the turbo is normal.
Don't be too worried about a bake out. They are necessary when you go down to 10-6 and beyond but they do not gain you much at 10-3 where a fusor runs. Pump on your chamber for a few hours or overnight and you will be in good shape. Once is is pumped out, keep it under vacuum. Keeping nitrogen or argon in the chamber will keep the water out but will not prevent gas molecules from diffusing back into the pores of the metal chamber. Dry gas is better than nothing but not as good as a tight chamber and a good vacuum.
Sounds like you are getting close to first light. If you have your power supply and grids ready, don't miss the opportunity to see what the plasma looks like as the vacuum comes down.
Frank S.
Don't be too worried about a bake out. They are necessary when you go down to 10-6 and beyond but they do not gain you much at 10-3 where a fusor runs. Pump on your chamber for a few hours or overnight and you will be in good shape. Once is is pumped out, keep it under vacuum. Keeping nitrogen or argon in the chamber will keep the water out but will not prevent gas molecules from diffusing back into the pores of the metal chamber. Dry gas is better than nothing but not as good as a tight chamber and a good vacuum.
Sounds like you are getting close to first light. If you have your power supply and grids ready, don't miss the opportunity to see what the plasma looks like as the vacuum comes down.
Frank S.
Achiever's madness; when enough is still not enough. ---FS
We have to stop looking at the world through our physical eyes. The universe is NOT what we see. It is the quantum world that is real. The rest is just an electron illusion. ---FS
We have to stop looking at the world through our physical eyes. The universe is NOT what we see. It is the quantum world that is real. The rest is just an electron illusion. ---FS
Re: Great American chamber bake-out
>Heating a vacuum chamber open to the air will not do a thing for you. It would be a waste of time.
This is pretty much what I assumed, but it doesn't hurt to get a second opinion!
>Don't be too worried about a bake out. They are necessary when you go down to 10-6 and beyond but they do not gain you much at 10-3 where a fusor runs. Pump on your chamber for a few hours or overnight and you will be in good shape.
Well, the reason I considered a bake-out was the high outgassing issue I've experinced. I recently tried to pump down using just a roughing pump (Welch 1402) for about 10 or 12 hours and with in 48 hours the pressure rose from 15 milltorr to 520 milltorr. I left it for another week and the pressure rose to about 540 milltorr. I tried spraying methanol all over the outside and I saw no pressure increases. so I am pretty sure it outgassing issue.
>Sounds like you are getting close to first light. If you have your power supply and grids ready, don't miss the opportunity to see what the plasma looks like as the vacuum comes down.
Not quite. I still have to purchase the Turbo pump, electrical feedthrus, finish construction of the grids and power supply. After stumbling on how much effort it takes to pump down the chamber, I am going to spend extra time constructing the grids and other internal workings. I am going with a multiple grid configuration so I can test several ideas with a single pump down.
Now I understand why every loaths having to open up the chamber to make changes. It takes damn forever pump down the chamber to a stable pressure!
Thanks,
This is pretty much what I assumed, but it doesn't hurt to get a second opinion!
>Don't be too worried about a bake out. They are necessary when you go down to 10-6 and beyond but they do not gain you much at 10-3 where a fusor runs. Pump on your chamber for a few hours or overnight and you will be in good shape.
Well, the reason I considered a bake-out was the high outgassing issue I've experinced. I recently tried to pump down using just a roughing pump (Welch 1402) for about 10 or 12 hours and with in 48 hours the pressure rose from 15 milltorr to 520 milltorr. I left it for another week and the pressure rose to about 540 milltorr. I tried spraying methanol all over the outside and I saw no pressure increases. so I am pretty sure it outgassing issue.
>Sounds like you are getting close to first light. If you have your power supply and grids ready, don't miss the opportunity to see what the plasma looks like as the vacuum comes down.
Not quite. I still have to purchase the Turbo pump, electrical feedthrus, finish construction of the grids and power supply. After stumbling on how much effort it takes to pump down the chamber, I am going to spend extra time constructing the grids and other internal workings. I am going with a multiple grid configuration so I can test several ideas with a single pump down.
Now I understand why every loaths having to open up the chamber to make changes. It takes damn forever pump down the chamber to a stable pressure!
Thanks,
Re: Great American chamber bake-out
This brings up a prime reason for having a HiVac valve on the fusor. So you can leave the Turbo running, while you change out things in the fusor. Then, all you have to do if rough out the fusor, and then open the highvac valve and your vacuum will be in usuable range in a minute or two.
If you were to put a small diameter (like 1/4" SS line with Swagelok fittings and small bellows sealed valve) around the High Vac valve, you could actually use the Turbo to pump directly, via the small diameter line. Then once the pressure is down, open the main valve... and you're ready for work.
If you intend to do fusion, you will need some sort of throttle valve and a needle valve to inlet the D2. So the bypass line is a good investment.
Dave Cooper
If you were to put a small diameter (like 1/4" SS line with Swagelok fittings and small bellows sealed valve) around the High Vac valve, you could actually use the Turbo to pump directly, via the small diameter line. Then once the pressure is down, open the main valve... and you're ready for work.
If you intend to do fusion, you will need some sort of throttle valve and a needle valve to inlet the D2. So the bypass line is a good investment.
Dave Cooper
- Richard Hull
- Moderator
- Posts: 15039
- Joined: Fri Jun 15, 2001 9:44 am
- Real name: Richard Hull
Re: Great American chamber bake-out
The key to getting right back into action, bypass line or not, is the rapidity of action in changing things in your chamber. Time and air moisture are very evil enemies to pumped system components once re-exposed to air.
Before opening a system as, John suggested, be sure to have every component and tool you will need at hand so that the work can proceed very rapidly and avoid at all costs touching the interior of the device with your bare hands, once opened. From seal break to reseal should be a matter of a couple of minutes at most.
If you take much longer to complete the task or if you run into a snag that drags on for tens of minutes, you will have a proportionally longer pump time. After an hour, consider the thing pretty fouled with water is you live in a moist climate. (East Coast).
Richard Hull
Before opening a system as, John suggested, be sure to have every component and tool you will need at hand so that the work can proceed very rapidly and avoid at all costs touching the interior of the device with your bare hands, once opened. From seal break to reseal should be a matter of a couple of minutes at most.
If you take much longer to complete the task or if you run into a snag that drags on for tens of minutes, you will have a proportionally longer pump time. After an hour, consider the thing pretty fouled with water is you live in a moist climate. (East Coast).
Richard Hull
Progress may have been a good thing once, but it just went on too long. - Yogi Berra
Fusion is the energy of the future....and it always will be
The more complex the idea put forward by the poor amateur, the more likely it will never see embodiment
Fusion is the energy of the future....and it always will be
The more complex the idea put forward by the poor amateur, the more likely it will never see embodiment
Re: Great American chamber bake-out
>Dave Cooper: If you intend to do fusion, you will need some sort of throttle valve and a needle valve to inlet the D2. So the bypass line is a good investment.
Thanks, I already have a UHV needle valve for D2 and I have a throttle baffle for the turbo on the way. My chamber has four 1.33 conflat ports three 2.75 ports, one 6 inch conflat and one 8 inch conflat..The 6 inch port will be used for the Turbo, the 8 inch will be reduced to a 6 inch and used for the HV feedthru. One 2.75 will be used for a view port, another for a 4 pole eletrical feedthru. One 1.33 will be used for the vacuum pressure sensor, another for the D2 input (attached to a needle valve). The unused ports will be blanked off.
However I am still trying to keep the number of connections to a mininum. I have two UHV 2.75 valves which I could used to attach to both the Turbo and the Rough pump (via 2.75 Tee). However I don't see any significant reason why I can't just rough pump down thru the Turbo output port. Is there a solid reason why I should not do this?
>Rich Hull: From seal break to reseal should be a matter of a couple of minutes at most.
I don't see that being possible. Speed and accuracy are pretty much mutually exclusive when it comes to my skills.
>Rich Hull: avoid at all costs touching the interior of the device with your bare hands, once opened.
Do you have a prefered brand of gloves you can recommend? I've been using latex and nitrile gloves when machining SS to avoid nasty SS swarf. But these gloves have a small amount of powder (even the "non-powdered" version). Is there a vendor that sells gloves that are "vacuum" certiified?
Thanks
Thanks, I already have a UHV needle valve for D2 and I have a throttle baffle for the turbo on the way. My chamber has four 1.33 conflat ports three 2.75 ports, one 6 inch conflat and one 8 inch conflat..The 6 inch port will be used for the Turbo, the 8 inch will be reduced to a 6 inch and used for the HV feedthru. One 2.75 will be used for a view port, another for a 4 pole eletrical feedthru. One 1.33 will be used for the vacuum pressure sensor, another for the D2 input (attached to a needle valve). The unused ports will be blanked off.
However I am still trying to keep the number of connections to a mininum. I have two UHV 2.75 valves which I could used to attach to both the Turbo and the Rough pump (via 2.75 Tee). However I don't see any significant reason why I can't just rough pump down thru the Turbo output port. Is there a solid reason why I should not do this?
>Rich Hull: From seal break to reseal should be a matter of a couple of minutes at most.
I don't see that being possible. Speed and accuracy are pretty much mutually exclusive when it comes to my skills.
>Rich Hull: avoid at all costs touching the interior of the device with your bare hands, once opened.
Do you have a prefered brand of gloves you can recommend? I've been using latex and nitrile gloves when machining SS to avoid nasty SS swarf. But these gloves have a small amount of powder (even the "non-powdered" version). Is there a vendor that sells gloves that are "vacuum" certiified?
Thanks
- Richard Hull
- Moderator
- Posts: 15039
- Joined: Fri Jun 15, 2001 9:44 am
- Real name: Richard Hull
Re: Great American chamber bake-out
When machining, I use no gloves and handle all items freely. I sort of expect a certain amount of cuts and scraps in this biz. I am far more concerned about something flying out of a chuck or out of a milling situation than a cut.
Once all machining is done all items are degreased according to vacuum standard practice.
Once assembled, evacuated and disassebly is mandated, all components are handled with clean, white cotton gloves.
Speed is easy during re-opening sessions with good planning in design and preparation prior to opening. Without speed or in the case of a major operation, a long re-pump session is just going to forever be in the cards.
Richard Hull
Once all machining is done all items are degreased according to vacuum standard practice.
Once assembled, evacuated and disassebly is mandated, all components are handled with clean, white cotton gloves.
Speed is easy during re-opening sessions with good planning in design and preparation prior to opening. Without speed or in the case of a major operation, a long re-pump session is just going to forever be in the cards.
Richard Hull
Progress may have been a good thing once, but it just went on too long. - Yogi Berra
Fusion is the energy of the future....and it always will be
The more complex the idea put forward by the poor amateur, the more likely it will never see embodiment
Fusion is the energy of the future....and it always will be
The more complex the idea put forward by the poor amateur, the more likely it will never see embodiment
Re: Great American chamber bake-out
A.G. -
You CAN rough through your Turbo, foreline connection, but this is really not desirable for at least two reasons: One is the large pieces dirt and trash, hazard. Many turbos have a screen to keep nuts, bolts and lost washers out of the impellers. Good investment for all to make.
But the real reason, for roughing through a separate line, is that the high pressure gas, is wet, possibly oily, but in general a lot more contaminated than the foreline gas will be after things get down to a few tens of microns. Some of that contamination will coat the interior of your Turbo and despite the flogging the impellers give to everything, their is still some boundary effects that partially screen the surfaces of the rotors and stators.
Some Varian manuals describe use of a clean gas at relatively high pressure...high for a turbo.. that is... as a way to scrub off the blades and interior parts. If the thing pumps down to the pressures you need and does it quickly, then it is clean enough, and the rest is just procdures for a dull Saturday, or something.
For High Vacuum, high voltage assembly, we have always used the latex powder free gloves, to minimize contamination.
Richard's caution about not touching with bare hands anything inside is the most important caution of all. The grease (everbody knows this don't they?) from your fingers is fairly volatile, and will be an outgassing source for literally days.
There used to be a poster about the vapor pressure of a fingerprint. (Varian or Welch? ) Once had one in the lab.
I am always leery of the cotton gloves, from the standpoint of lint, but there is a nearly lint free type. And their softness is a plus on polished surfaces. Matter of choice, here, I think.
One thing to remember about the rubber gloves, is that they can easily become contaminated on the outside, with grease oil, donuts, coffee, and other common lab "materials".
A lot of opitcal folks, like the finger cots. I don't, but they do let the rest of your hands breathe while work their damage on the circulation to your fingers. They probably are not the best for vacuum assembly as your palms are not covered.
Sounds like you have your gas inlet metering system well in hand.
Dave Cooper
You CAN rough through your Turbo, foreline connection, but this is really not desirable for at least two reasons: One is the large pieces dirt and trash, hazard. Many turbos have a screen to keep nuts, bolts and lost washers out of the impellers. Good investment for all to make.
But the real reason, for roughing through a separate line, is that the high pressure gas, is wet, possibly oily, but in general a lot more contaminated than the foreline gas will be after things get down to a few tens of microns. Some of that contamination will coat the interior of your Turbo and despite the flogging the impellers give to everything, their is still some boundary effects that partially screen the surfaces of the rotors and stators.
Some Varian manuals describe use of a clean gas at relatively high pressure...high for a turbo.. that is... as a way to scrub off the blades and interior parts. If the thing pumps down to the pressures you need and does it quickly, then it is clean enough, and the rest is just procdures for a dull Saturday, or something.
For High Vacuum, high voltage assembly, we have always used the latex powder free gloves, to minimize contamination.
Richard's caution about not touching with bare hands anything inside is the most important caution of all. The grease (everbody knows this don't they?) from your fingers is fairly volatile, and will be an outgassing source for literally days.
There used to be a poster about the vapor pressure of a fingerprint. (Varian or Welch? ) Once had one in the lab.
I am always leery of the cotton gloves, from the standpoint of lint, but there is a nearly lint free type. And their softness is a plus on polished surfaces. Matter of choice, here, I think.
One thing to remember about the rubber gloves, is that they can easily become contaminated on the outside, with grease oil, donuts, coffee, and other common lab "materials".
A lot of opitcal folks, like the finger cots. I don't, but they do let the rest of your hands breathe while work their damage on the circulation to your fingers. They probably are not the best for vacuum assembly as your palms are not covered.
Sounds like you have your gas inlet metering system well in hand.
Dave Cooper
-
- Posts: 443
- Joined: Thu Mar 11, 2004 12:38 am
- Real name: Todd Massure
Re: Great American chamber bake-out
Could I expect to get less outgassing from the ceramic parts in a fusor than the metal per given area? Does the feedthrough require any different preparation and handling than the metal?
-Todd
Re: Great American chamber bake-out
>When machining, I use no gloves and handle all items freely. I sort of expect a certain amount of cuts and scraps in this biz. I am far more concerned about something flying out of a chuck or out of a milling situation than a cut.
I use gloves because I usually ended up with a half a dozen tiny, razor sharp stainless splinters when machining with stainless. The gloves definately help. The splinters are often extremely difficult to remove or even see if your hands are dirty.
I use gloves because I usually ended up with a half a dozen tiny, razor sharp stainless splinters when machining with stainless. The gloves definately help. The splinters are often extremely difficult to remove or even see if your hands are dirty.
Re: Great American chamber bake-out
>Some of that contamination will coat the interior of your Turbo and despite the flogging the impellers give to everything, their is still some boundary effects that partially screen the surfaces of the rotors and stators.
I hadn't thought about that. That's sounds like a good point
>For High Vacuum, high voltage assembly, we have always used the latex powder free gloves, to minimize contamination.
All the brands of "powder free" latex seem to have some powder. The best I came across are "Nitrile" gloves which don't have any powder. I was hoping you might know of a vendor that sells latex for clean rooms.
What about latex and acetone? I was concerned that acetone might react with the acetone. From what I read, Acetone and Methanol are the two best for degreasing vacuum devices since they leave no residue. However, I really don't like the idea of using Methanol for cleaning since its very toxic and can seep past skin into the bloodstream.
Thanks
I hadn't thought about that. That's sounds like a good point
>For High Vacuum, high voltage assembly, we have always used the latex powder free gloves, to minimize contamination.
All the brands of "powder free" latex seem to have some powder. The best I came across are "Nitrile" gloves which don't have any powder. I was hoping you might know of a vendor that sells latex for clean rooms.
What about latex and acetone? I was concerned that acetone might react with the acetone. From what I read, Acetone and Methanol are the two best for degreasing vacuum devices since they leave no residue. However, I really don't like the idea of using Methanol for cleaning since its very toxic and can seep past skin into the bloodstream.
Thanks
-
- Posts: 443
- Joined: Thu Mar 11, 2004 12:38 am
- Real name: Todd Massure
Re: Great American chamber bake-out
I seem to remember that the cheapo vinyl gloves like food handlers use didn't seem to have any powder
Re: Great American chamber bake-out
>Could I expect to get less outgassing from the ceramic parts in a fusor than the metal per given area? Does the feedthrough require any different preparation and handling than the metal?
To my knowledge, as long as your not using a metal containing lead, cadium or zinc (galvanized, brass) you should be fine. The problem is that all metals can retain water and hydrogen on the surface. When you pump down these substances (and any containments) will outgas.
One issue you may run into with ceramics is with porous materals that can hold water, air or other containments that will outgas. You should not have a problem with the cermics from manfactured feedthru. However if you plan to place cermics inside, be sure to used alumina or glasses that are not porous. I believe some of the glazes used on ceramic can also be a source of outgassing.
To my knowledge, as long as your not using a metal containing lead, cadium or zinc (galvanized, brass) you should be fine. The problem is that all metals can retain water and hydrogen on the surface. When you pump down these substances (and any containments) will outgas.
One issue you may run into with ceramics is with porous materals that can hold water, air or other containments that will outgas. You should not have a problem with the cermics from manfactured feedthru. However if you plan to place cermics inside, be sure to used alumina or glasses that are not porous. I believe some of the glazes used on ceramic can also be a source of outgassing.
Re: Great American chamber bake-out
Sorry about forgetting the glove source. One type is medical latex gloves which are powder free. Brand is Evolution, but not sure of supplier. Almost certain McMaster -Carr carries them, would think other suppliers like Cole-Parmer, VWR, and Wal-Mart carries them.. (but not sure if they are powder-free).
Acetone and methanol will attack them in time. I always use them if I have to clean things extensively with these solvents.
Using these solvents without the gloves, often makes it very difficult if not impossible to avoid depositing finger grease which the solvents extract from your skin.
Hope that's some help.
Dave Cooper
Acetone and methanol will attack them in time. I always use them if I have to clean things extensively with these solvents.
Using these solvents without the gloves, often makes it very difficult if not impossible to avoid depositing finger grease which the solvents extract from your skin.
Hope that's some help.
Dave Cooper
-
- Posts: 1494
- Joined: Thu Jun 28, 2001 1:30 am
- Real name: Jon Rosenstiel
- Location: Southern California
Re: Great American chamber bake-out
Anonymous Guy wrote:
> I use gloves because I usually ended up with a half a dozen tiny, razor sharp stainless splinters when machining with stainless. The gloves definately help. The splinters are often extremely difficult to remove or even see if your hands are dirty.
Please be careful. Wearing gloves around machinery is not a good idea. Much better to have a few cuts or splinters than to have your extremities pulled into a machine.
Jon Rosenstiel
> I use gloves because I usually ended up with a half a dozen tiny, razor sharp stainless splinters when machining with stainless. The gloves definately help. The splinters are often extremely difficult to remove or even see if your hands are dirty.
Please be careful. Wearing gloves around machinery is not a good idea. Much better to have a few cuts or splinters than to have your extremities pulled into a machine.
Jon Rosenstiel
- Richard Hull
- Moderator
- Posts: 15039
- Joined: Fri Jun 15, 2001 9:44 am
- Real name: Richard Hull
Re: Great American chamber bake-out
Jon is right, I have had a couple of close calls over the years and one horrible accident where a lathedog hit my hand and knocked it not into the machine, but back around behind me. Hurt like a son- of-a-bitch and bled like a stuck pig.
Rolledup long sleeve shirts are a no-no as they will unroll at a critical moment and pull an arm into the works. The classic light weight leather lathe apron is a real nice thing to have. Bare hands and a lexan face shield are best. Bare hands are best protected by filing or deburring rough or sharp machined edges prior to removal of chucked work. Goggles are fine, too, but the face shield will help deflect a small missle or two that goggles will not.
Still, you will get scrapes and cuts no matter what. The number, severity and frequency are directly proportional to your total machining time over the years and your attention to deburring sharp fragments prior to handling.
Richard Hull
Rolledup long sleeve shirts are a no-no as they will unroll at a critical moment and pull an arm into the works. The classic light weight leather lathe apron is a real nice thing to have. Bare hands and a lexan face shield are best. Bare hands are best protected by filing or deburring rough or sharp machined edges prior to removal of chucked work. Goggles are fine, too, but the face shield will help deflect a small missle or two that goggles will not.
Still, you will get scrapes and cuts no matter what. The number, severity and frequency are directly proportional to your total machining time over the years and your attention to deburring sharp fragments prior to handling.
Richard Hull
Progress may have been a good thing once, but it just went on too long. - Yogi Berra
Fusion is the energy of the future....and it always will be
The more complex the idea put forward by the poor amateur, the more likely it will never see embodiment
Fusion is the energy of the future....and it always will be
The more complex the idea put forward by the poor amateur, the more likely it will never see embodiment
Re: Great American chamber bake-out
>Please be careful. Wearing gloves around machinery is not a good idea. Much better to have a few cuts or splinters than to have your extremities pulled into a machine.
Nothing to worry about. I am wearing latex gloves, similar to surgical latex gloves. If the glove got caught in the machinary, it would break. They are very thin and break pretty easy and are designed to use with machinery. I would never consider wearing cotton or work gloves.
Nothing to worry about. I am wearing latex gloves, similar to surgical latex gloves. If the glove got caught in the machinary, it would break. They are very thin and break pretty easy and are designed to use with machinery. I would never consider wearing cotton or work gloves.
Re: Great American chamber bake-out
>Using these solvents without the gloves, often makes it very difficult if not impossible to avoid depositing finger grease which the solvents extract from your skin.
What I was concerned is that the solvent would break down the latex and leave a latex reside on metal surface and present another source of outgassing. But, if you haven't had any trouble, I will consider using them.
>One type is medical latex gloves which are powder free. Brand is Evolution, but not sure of supplier
Thanks, I didn't think about using surgical grade gloves, but that makes sense. Any containments on gloves would likely end up in the patient, so that must be very clean.
What I was concerned is that the solvent would break down the latex and leave a latex reside on metal surface and present another source of outgassing. But, if you haven't had any trouble, I will consider using them.
>One type is medical latex gloves which are powder free. Brand is Evolution, but not sure of supplier
Thanks, I didn't think about using surgical grade gloves, but that makes sense. Any containments on gloves would likely end up in the patient, so that must be very clean.
Re: Great American chamber bake-out
>Bare hands are best protected by filing or deburring rough or sharp machined edges prior to removal of chucked work
The problem isn't the machined edges is the taillings left from the machining mixed in with the lubercant. They end up every where. Its impossible to clean them up with out getting some on your hands, and they are very tiny and razor sharp. It doesn't take much pressure to get a few lodged under your skin.
>have had a couple of close calls over the years and one horrible accident where a lathedog hit my hand and knocked it not into the machine, but back around behind me. Hurt like a son- of-a-bitch and bled like a stuck pig.
I haven't done any work on a lathe yet, just a vertical mill. For my turned parts I use a radial table or a rotary indexer with a lathe chuck. I would like to get a lathe, but it would be real a pain to get it moved into the basement.
When I operate the machine I am very careful to keep all body parts and clothing away from spinning tools. So far I've not had any accidents, except perhaps hitting the back of my hand on a stationary tool bit while tighting down a clamp with a wrench. Carbide bits are extremely sharp!
Another import issue to to avoid touching your head where swarf (tailings) can get into your eyes, even if you wear googles or a face shield. I am always concerned that tiny piece will end up in an eye or even on a lip where it can be ingested. Another issue is tracking swarf upstairs since tiny sharp pieces can get stuck on the bottom of your shoes or even your clothing, and end up on the floor. Stainless is awsome metal, but machining it absolutely sucks. I much perfer to machine aluminum or even steel.
The problem isn't the machined edges is the taillings left from the machining mixed in with the lubercant. They end up every where. Its impossible to clean them up with out getting some on your hands, and they are very tiny and razor sharp. It doesn't take much pressure to get a few lodged under your skin.
>have had a couple of close calls over the years and one horrible accident where a lathedog hit my hand and knocked it not into the machine, but back around behind me. Hurt like a son- of-a-bitch and bled like a stuck pig.
I haven't done any work on a lathe yet, just a vertical mill. For my turned parts I use a radial table or a rotary indexer with a lathe chuck. I would like to get a lathe, but it would be real a pain to get it moved into the basement.
When I operate the machine I am very careful to keep all body parts and clothing away from spinning tools. So far I've not had any accidents, except perhaps hitting the back of my hand on a stationary tool bit while tighting down a clamp with a wrench. Carbide bits are extremely sharp!
Another import issue to to avoid touching your head where swarf (tailings) can get into your eyes, even if you wear googles or a face shield. I am always concerned that tiny piece will end up in an eye or even on a lip where it can be ingested. Another issue is tracking swarf upstairs since tiny sharp pieces can get stuck on the bottom of your shoes or even your clothing, and end up on the floor. Stainless is awsome metal, but machining it absolutely sucks. I much perfer to machine aluminum or even steel.
- Richard Hull
- Moderator
- Posts: 15039
- Joined: Fri Jun 15, 2001 9:44 am
- Real name: Richard Hull
Re: Great American chamber bake-out
Machining stock 304 stainless is actually easier than machining mild steel. The key is to use the absolute correct computed tool speed. It cuts similar to brass and carbide tools are not needed, though they help a bit by staying sharp longer. As usual most of the bad rep on stainless comes from folks burning or chattering tools to death without taking the time to compute and adjust the machine speed to precisely the recommended turning surface speed. Once done, I have done light SS machining DRY! It has to be seen to be believed by all who have smoked or chipped carbide tools on SS machining.
Richard Hull
Richard Hull
Progress may have been a good thing once, but it just went on too long. - Yogi Berra
Fusion is the energy of the future....and it always will be
The more complex the idea put forward by the poor amateur, the more likely it will never see embodiment
Fusion is the energy of the future....and it always will be
The more complex the idea put forward by the poor amateur, the more likely it will never see embodiment
Re: Great American chamber bake-out
>Machining stock 304 stainless is actually easier than machining mild steel. The key is to use the absolute correct computed tool speed. It cuts similar to brass and carbide tools are not needed, though they help a bit by staying sharp longer.
Cutting Stainless isn't the problem, Its the super sharp and tiny chips that I hate.
Cutting Stainless isn't the problem, Its the super sharp and tiny chips that I hate.
Re: Great American chamber bake-out
Hi AG:
Fusion wants your BLOOD!
Happy Halloween!
Larry Leins
Fusor Tech
Fusion wants your BLOOD!
Happy Halloween!
Larry Leins
Fusor Tech