For Short Term Learning Discussions ONLY. This area is for CURSORY questions and connecting with other users ONLY. ALL technical contributions need to be made in the appropriate forums and NOT HERE. All posts are temporary and will be deleted within weeks or months. You should have already search the extensive FAQs in each of the forums before posting here as your question may already be answered.
Peter Schmelcher
Posts: 228
Joined: Sat Nov 13, 2010 1:56 am
Real name: Peter Schmelcher

Re: Is this neutron counter dead?

Post by Peter Schmelcher »

Your operating voltage seems high to me. I tested a CH19N and determined it was working by ramping the voltage up in steps and measuring the resulting pulse heights (Geiger plateau curve). Pulse height is the gain of the detection process and increases with voltage.
From my memory I believe the best signal to noise was around 1200 +-75V for proportional operation. The manufacturer recommended voltage was 1875V to 2100V for Geiger operation but honestly it was unusable at that voltage.
At higher tube operating voltages the circuit you are using becomes an avalanche oscillator.
Cheers -Peter