by Richard Hull » Wed Jul 30, 2014 5:45 am
I have some questions and have used the GE tube you have before and the bias voltage is absolutely critical and a normal back ground for the GE tube should be very, very low, maybe under 1 or 2 CPM, at most! I fear you might be operating this tube in a geiger mode and all your count data is due to x rays!!
The proper bias on these old tubes can have a wide range from 500 volts to 700 volts. The actual useful span, once the useful bias range is found, is very narrow, making this tube a bear to find its "gamma quiet zone". Much higher than 600 or 700 volts bias and the background goes up significantly due to NORM and other noise related issues. I see the meter reading in your image is about 900 volts! Way, way too high for the tube!! Even a superb GM counter will not normally exceed 50 CPM background. Something is very wrong with your numbers. It may be a bad tube, but the voltage is far too high to allow proper neutron counting. I suspect you are x-ray counting. Yes, you may be counting neutrons as well, but no way will the GE tube count that high a background with correct bias.
I am rather shocked no one else caught this. Perhaps no one here who has used the GE tube has seen this posting. You are up against folks who know this tube and its normal characteristics. No gas counter tube in the world will have a background of over 50-70 CPM when setup correctly. Correctly adjusted, normal neutron counters of the BF3 and 3He types never have a background of over 5-10 CPM which are mostly huge cosmic particle star sprays and ground based neutrons. The GE tube is far less sensitive and 1-3 CPM would be a norm.
Not nearly enough data to be included in the neutron club. Make sure you review the neutron club rules.
Reset your bias to a 2 or 3 CPM background level and then re-run your tests You may find that at your operational level the count might increase by only 15-20 additonal CPM over background, maybe more. The big test, once the tube is set properly, is to get the increased count, (making neutrons), then with the machine still running, remove the moderator far from the tube and the count should go back to background levels as it it is not now seeing neutrons anymore. If the count does not go down or goes up, you are still just seeing x-rays and no neutrons are present, you are not doing fusion. It is important to get the tube out of the moderator. tubes should be very loose in any moderator. A 1 inch diameter tube will work just fine in a 2" hole in the moderator. having a moderator hug the tube tightly is just not needed at all and gains nothing, in the end.
Again these tubes can be a bear to set up, but once setup correctly, they are fine tubes and are moderately sensitive to neutrons in a good moderator placed close to the fusor chamber.