Silicon Avalanche Photodiodes and Scintillation Crystals: SiAPD Redux

Post by Edward Miller »

There's an eBay seller offering Perkin Elmer SiAPDs at $91 each. I really want to move from the big/fragile PMTs to APDs. Does anyone have experience with APD circuits? Let's collaborate on APDs.

http://www.ebay.com/itm/Photo-Diode-Sil ... 2ec0f3c05b
User avatar
Steven Sesselmann
Posts: 2128
Joined: Wed Aug 10, 2005 9:50 pm
Real name: Steven Sesselmann
Location: Sydney - Australia
Contact:

Re: Silicon Avalanche Photodiodes and Scintillation Crystals: SiAPD Redux

Post by Steven Sesselmann »

Ed,

Looks interesting, could these work with Scintillators to give pulse height?

Steven
http://www.gammaspectacular.com - Gamma Spectrometry Systems
https://www.researchgate.net/profile/Steven_Sesselmann - Various papers and patents on RG
John Futter
Posts: 1850
Joined: Wed Apr 21, 2004 10:29 pm
Real name: John Futter
Contact:

Re: Silicon Avalanche Photodiodes and Scintillation Crystals: SiAPD Redux

Post by John Futter »

Edward
Just doing some mental gymnastics on the noise --this would be equivalent to using a PMT without a preamp ie short by a factor of 1000 to 10000 in sensitivity / noise performance.
For nuclear use it is possible to use photopin diodes / avalanche diodes natively ie reversed biased and using the particle to produce a charge cascade in the device. you will also need a very good fast charge sensitive preamp for this
downside is that it only works for reasonably weak energy particles as the silicon has little stopping power --good from 100's of eV to low 10,000s of eV.
Another downside is that the particles permanently damage the junction and leakage rises with damage finally making them useless for spectroscopic work.
Better diodes for this are surface barrier detector diodes but they suffer the same damage mechanism but have much better energy resolution