BNC cables, terminators, Tees, etc.

This area is for discussions involving any fusion related radiation metrology issues. Neutrons are the key signature of fusion, but other radiations are of interest to the amateur fusioneer as well.
Post Reply
Tom Dressel
Posts: 287
Joined: Fri Jun 22, 2001 4:44 pm
Real name:

Re: BNC cables, terminators, Tees, etc.

Post by Tom Dressel »

I talked again to the sales depertment at Bertan again today, reguarding the power supply cables. As stated previously, the weak link in the system is the standard BNC connector not the RG-58 cable. The RG-58 cable is tested by Bertan to 12 KV DC. The Kings 1064-1 "10 KV" connector is tested to 15 KV DC. So if you are going to run the cable and connector over ~500 volts, it should be terminated with a MHV or preferably a SHV connector, again this is due to the creep path limitations of the BNC connectors.

Tom Dressel