Would it be possible to run two diffusion pumps in parallel if properly controlled? Say each pump with a inlet valve both connected to a cold trap and then to the vacuum chamber?
This is for a large vacuum chamber and these pumps are what is available.
Parallel diffusion pumps.
-
- Posts: 10
- Joined: Fri Jan 08, 2021 6:02 pm
- Real name: Marko Kuzmanovic
- Location: Auckland
- Richard Hull
- Moderator
- Posts: 15037
- Joined: Fri Jun 15, 2001 9:44 am
- Real name: Richard Hull
Re: Parallel diffusion pumps.
The key is with proper operation control and plumbing, I would imagine parallel pumping is possible. I would never try it, but it sounds like your back is against the wall.
Richard Hull
Richard Hull
Progress may have been a good thing once, but it just went on too long. - Yogi Berra
Fusion is the energy of the future....and it always will be
The more complex the idea put forward by the poor amateur, the more likely it will never see embodiment
Fusion is the energy of the future....and it always will be
The more complex the idea put forward by the poor amateur, the more likely it will never see embodiment
- Maciek Szymanski
- Posts: 193
- Joined: Wed Nov 20, 2019 2:31 pm
- Real name: Maciek Szymański
- Location: Warsaw, Poland
Re: Parallel diffusion pumps.
One of the large four beam interaction chambers at IPPLM was pumped by two 2000 l/s diffusion pumps in parallel:
The arrow indicates the first pump flange with the forevacuum bypass valve. The second pump is beyond the picure frame connected to the same large tube. I even still have those 500 mm manifolds in the stockroom at work. When I talked with Leybold representative he told me that they just set up a vacuum system for industrial glass plating using not less than ten 3000 l/s turbos in parallel...
So definitely possible. For other hand you must have a REALLY big chamber or an enormous leak. In the laser lab we have the optical compressor chamber of more than 1 m3 volume which is pumped by a single 300 l/s turbo backed with a small scroll. It gets down to 1e-6 mbar in 30 to 40 minutes. The most of this time is pre evacuation by the scroll pump.
The arrow indicates the first pump flange with the forevacuum bypass valve. The second pump is beyond the picure frame connected to the same large tube. I even still have those 500 mm manifolds in the stockroom at work. When I talked with Leybold representative he told me that they just set up a vacuum system for industrial glass plating using not less than ten 3000 l/s turbos in parallel...
So definitely possible. For other hand you must have a REALLY big chamber or an enormous leak. In the laser lab we have the optical compressor chamber of more than 1 m3 volume which is pumped by a single 300 l/s turbo backed with a small scroll. It gets down to 1e-6 mbar in 30 to 40 minutes. The most of this time is pre evacuation by the scroll pump.
“Begin at the beginning," the King said, very gravely, "and go on till you come to the end: then stop.” ― Lewis Carroll, Alice in Wonderland
-
- Posts: 131
- Joined: Sat Apr 11, 2020 9:47 am
- Real name: Christoffer Braestrup
Re: Parallel diffusion pumps.
There is almost no upper limit to the size of diffusion pumps, wouldn't one larger pump usually make more sense?
If one "needs" a large leak - it's common to use multiple pumps in parallel, but having small orifices between their sectors of the vacuum chamber, a so-called differentially pumped system. This is often done on mass spectrometers and beam lines.
I'm curious about your application!
If one "needs" a large leak - it's common to use multiple pumps in parallel, but having small orifices between their sectors of the vacuum chamber, a so-called differentially pumped system. This is often done on mass spectrometers and beam lines.
I'm curious about your application!
- Maciek Szymanski
- Posts: 193
- Joined: Wed Nov 20, 2019 2:31 pm
- Real name: Maciek Szymański
- Location: Warsaw, Poland
Re: Parallel diffusion pumps.
You can build huge diff pumps, but nobody can stop you from connecting few of those monsters in parallel to even bigger chamber. Like this Goddard Space Flight Center test chamber:
(Sorry, can't stop myself)
Sometimes it makes sense to have more than one pump in parallel. It is possible to use just one pump to just maintain the vacuum, or switch the additional pump only, when there is lagre gas load (like heavily outgassing material in the chamber, pulsed gas jet etc.) and it's not possible to use the differential pumping. Oversize pump will use adequate high power heater which will be heating up the laboratory even when the full capacity is not needed. It can be a real problem in the laser lab.
(Sorry, can't stop myself)
Sometimes it makes sense to have more than one pump in parallel. It is possible to use just one pump to just maintain the vacuum, or switch the additional pump only, when there is lagre gas load (like heavily outgassing material in the chamber, pulsed gas jet etc.) and it's not possible to use the differential pumping. Oversize pump will use adequate high power heater which will be heating up the laboratory even when the full capacity is not needed. It can be a real problem in the laser lab.