This is yet another demonstration of why fusor chambers should not be made of glass or other brittle materials. It may also serve as a warning for those building small chambers based on 2.75" conflat systems.
I have a 2.75" conflat 6-way chamber with two viewports, one on axis with a graphite and tungsten cylindrical grid and the other perpendicular to it. The grid is oriented such that ion beams emanating from the gaps hit stainless steel, but in my recent fusion runs a problem has arisen.
I use a threaded rod to connect the grid (the graphite is threaded) to the feedthrough and the assembly hangs upside down. After I pumped the chamber down the grid rotated slightly, but to a position where another grid wire was positioned before the viewport. It was a warning of things to come. Early on several ion beams were visible on camera but these eventually reduced to two very bright beams on axis with the viewport. The grid had rotated at least three times was about 2mm further from the insulator stalk. I suspect thermal cycling is the culprit. Since graphite is self-lubricating it takes very, very little force to move the grid in the direction gravity likes. I did not, however, realize that the grid had rotated this much while I was fusing. Everything is covered in a lead vest and the camera (of poor quality) is on axis with the grid, so there was no way to tell where the wires were.
Here comes the failure.
After about an hour and a half of fusing and conditioning the chamber I heard a loud pop, not unlike some short capacitive arc discharge. I've been having some arcing difficulties, but the pressure gauge told a different story, as did the smell of burning plastic. I quickly tripped the HV, isolated the turbo pump, and poked around the fusor until I removed the lead vest, revealing the culprit. The off-axis viewport was badly cracked and had luckily not imploded. The turbo pump would not have liked that.
The cracked viewport right after failure and still bolted to the chamber.
The chamber was ~100C but the viewport was much hotter, my guess was >150C. By the time I could get an IR thermometer everything had cooled down significantly. Upon closer inspection, I realized the gap between the grid wires pointed right to the epicenter of failure. The ion beam cooked the viewport and caused significant thermal stress until it finally gave up the ghost. My lead vest burned exclusively at the ion beam spot despite being in direct contact with all the glass and the flange. It is a testament to the locality of the heating.
Grid alignment looking in from viewport flange. The threaded rod is also barely visible.
Had the chamber been fully glass the grid spinning out of alignment would not have mattered. A bell jar would stand little chance, especially at high power since I was running at <100W. The failure could have been ex(im)plosive, but I got off lucky. Sacrificial glass is on order and grid construction is also up for revision. While my ion beam was uncannily focused, people with small chambers based on 2.75" crosses, tees, etc... might want to consider sacrificial glass since the viewport is so close to the action. Magnet deflectors (look up Andrew Seltzman's work) might also do the trick.
A closer look.
-Liam David
I have a 2.75" conflat 6-way chamber with two viewports, one on axis with a graphite and tungsten cylindrical grid and the other perpendicular to it. The grid is oriented such that ion beams emanating from the gaps hit stainless steel, but in my recent fusion runs a problem has arisen.
I use a threaded rod to connect the grid (the graphite is threaded) to the feedthrough and the assembly hangs upside down. After I pumped the chamber down the grid rotated slightly, but to a position where another grid wire was positioned before the viewport. It was a warning of things to come. Early on several ion beams were visible on camera but these eventually reduced to two very bright beams on axis with the viewport. The grid had rotated at least three times was about 2mm further from the insulator stalk. I suspect thermal cycling is the culprit. Since graphite is self-lubricating it takes very, very little force to move the grid in the direction gravity likes. I did not, however, realize that the grid had rotated this much while I was fusing. Everything is covered in a lead vest and the camera (of poor quality) is on axis with the grid, so there was no way to tell where the wires were.
Here comes the failure.
After about an hour and a half of fusing and conditioning the chamber I heard a loud pop, not unlike some short capacitive arc discharge. I've been having some arcing difficulties, but the pressure gauge told a different story, as did the smell of burning plastic. I quickly tripped the HV, isolated the turbo pump, and poked around the fusor until I removed the lead vest, revealing the culprit. The off-axis viewport was badly cracked and had luckily not imploded. The turbo pump would not have liked that.
The cracked viewport right after failure and still bolted to the chamber.
The chamber was ~100C but the viewport was much hotter, my guess was >150C. By the time I could get an IR thermometer everything had cooled down significantly. Upon closer inspection, I realized the gap between the grid wires pointed right to the epicenter of failure. The ion beam cooked the viewport and caused significant thermal stress until it finally gave up the ghost. My lead vest burned exclusively at the ion beam spot despite being in direct contact with all the glass and the flange. It is a testament to the locality of the heating.
Grid alignment looking in from viewport flange. The threaded rod is also barely visible.
Had the chamber been fully glass the grid spinning out of alignment would not have mattered. A bell jar would stand little chance, especially at high power since I was running at <100W. The failure could have been ex(im)plosive, but I got off lucky. Sacrificial glass is on order and grid construction is also up for revision. While my ion beam was uncannily focused, people with small chambers based on 2.75" crosses, tees, etc... might want to consider sacrificial glass since the viewport is so close to the action. Magnet deflectors (look up Andrew Seltzman's work) might also do the trick.
A closer look.
-Liam David